Four ports measurements #350
Replies: 3 comments
-
|
Hi Nicolas, this annoyed me as well and I made some improvements to the Touchstone export a while ago. But they are not in the 1.6.0 version you are using. Please try updating to 1.6.4 (or even better: use the latest development version from here). 1.6.4 will preselect the most fitting traces when you export Touchstone files. The development version will also preselect the traces but if you then go and make changes to these selections, it will remember those changes when you open the export dialog again.
Yes, you can connect the reference output of VNA 1 to the reference input of VNA 2. You then also need to set the reference out setting to 10 MHz and the reference in setting to auto. This will result in VNA 2 getting its clock from VNA 1. I recommend terminating the unused reference input on VNA 1, otherwise it might pick up noise and try to use that as a reference (because it will also be set for auto-detecting the reference input signal) |
Beta Was this translation helpful? Give feedback.
-
|
Hi Jankae, log1_with 10MHz from VNA1.txt |
Beta Was this translation helpful? Give feedback.
-
Beta Was this translation helpful? Give feedback.




Uh oh!
There was an error while loading. Please reload this page.
-
Hi,
I'am trying 4 ports measurement (LibreVNA Version (64 bit): 1.6.0-b4a22176a - OS: Windows 10 Version 2009 - CPU Arch: x86_64).
Is it possible to automatically set the traces selection for export of 4 ports touchstone format to the right assignment? By default, everyone are at "none", and do it for the sixteen parameters each time we start the export menu is annoying.
Maybe to be sure to use the same reference frequency between LibreVNA, without an external clock, is it possible to "daisy chain" the ref freq of the first VNA (for instance carrying S11) out to In of the second VNA?
Thank you.
Nicolas.